(Piezoresistive Pressure Sensors / Blood Pressure Packages, ¾ÐÀúÇ×½Ä ¾Ð·Â¼¾¼)
The BP Series is ideal for high-volume, disposable medical applications and is compliant with the AAMI BP22 specification for blood pressure monitoring.
Overview
Features
- Pressure range of -30 to 300 mmHg
- Excellent Burst Pressure
- Compliant with AAMI BP22 specifications
- Dielectric gel barrier
- Fully Tested
- Small, disposable, low-cost package
- Temperature Compensated
- Solid state piezoresistive sensor
- Top side pressure entry
- Easy to install into customer¡¯s disposable blood pressure device housing
- Compatible with automated assembly equipment
- Compatible in air, gas and liquid
- Shipped as 120UP snapstrates or on tape and reel
- Automated testing and assembly
- Custom Options Available
- Engineering, design and manufacturing under one roof in Utah, US.
Features
- Disposable Blood Pressure measurement
- Kidney Dialysis Machines
- Infusion Pumps
- Surgical Procedures
Applications
Merit Sensor Systems, Inc. has partnered with customers for more than 20 years to design, fabricate, assemble and package reliable, cost-effective piezoresistive pressure sensor solutions for automotive, medical, industrial, aviation, defense and consumer applications.
Merit Sensor offers full-service design capabilities, in-house wafer fabrication, flexible shipping, packaging and assembly, piezoresistive technology (PRT), expansive pressure ranges (0.15 psi to 15,000 psi), complete pressure measurement (absolute, gage, differential and vacuum) and unmatched operating temperature range (-40¡ÆC to 150¡ÆC) via our Sentium¢ç process. Additionally, Merit Sensor is able to provide unparalleled flexibility to customize pressure sensing solutions to fit into our customers¡¯ applications.
Unlike other pressure sensor suppliers, Merit Sensor can provide customers with completely customized pressure sensor designs with large or smaller/limited production runs. Our customers range from pressure sensor transducer manufacturers who are already experts in pressure sensing technology and rely on Merit Sensor for highly stable and sensitive MEMS sensing elements (bare die), to customers who have little to no experience in the pressure sensing world and look to Merit Sensor to assist with a completely custom design and implementation of a pressure sensor that best fits their application.
At Merit Sensor our engineers are application experts. We are ready to help customers design your application to work with a pressure sensor, and/or design a customized pressure sensing solution that works for your application. If you are unsure whether a pressure sensor is right for your application, Merit Sensor can help you make that determination.
Specifications
|
Minimum |
Typical |
Maximum |
Units |
Notes |
General |
Pressure Range |
-30 |
¡¡ |
300 |
mmHg |
¡¡ |
Overpressure |
125 |
¡¡ |
¡¡ |
PSI |
¡¡ |
Electrical (22¡ÆC unless otherwise stated) |
Input Excitation |
1 |
¡¡ |
10 |
VDC |
Calibrated to 6 VDC |
Risk Current |
¡¡ |
¡¡ |
2 |
µA |
Previously qualified, not tested in production |
Input Impedance |
1,200 |
¡¡ |
3,000 |
¥Ø |
¡¡ |
Output Impedance |
285 |
¡¡ |
315 |
¥Ø |
¡¡ |
Dielectric Breakdown |
10,000 |
¡¡ |
VDC |
¡¡ |
Previously qualified, not tested in production |
Environmental |
Temp (Comp/Operating) |
15 |
¡¡ |
40 |
¡ÆC |
¡¡ |
Temperature (Storage) |
-25 |
¡¡ |
70 |
¡ÆC |
¡¡ |
Humidity |
10 |
¡¡ |
90 |
% |
¡¡ |
Light Sensitivity |
¡¡ |
¡¡ |
1 |
mmHg |
Previously qualified, not tested in production â¢æ¡° Per AAMI BP22 |
Operating Product Life |
168 |
¡¡ |
¡¡ |
Hours |
¡¡ |
Shelf Life |
3 |
¡¡ |
¡¡ |
Years |
¡¡ |
Sterilization (ETO) |
3 |
¡¡ |
¡¡ |
Cycles |
Sterilization performed by customer |
Mechanical |
Weight |
¡¡ |
¡¡ |
2 |
Grams |
¡¡ |
Volume Displacement |
¡¡ |
¡¡ |
0.02 |
mm3 |
¡¡ |
Media Interface |
¡¡ |
¡¡ |
¡¡ |
¡¡ |
Medical grade, dielectric gel |
Gel Tube Interface Material |
¡¡ |
¡¡ |
¡¡ |
¡¡ |
Polycarbonate |
Performance |
Assumes 6 VDC, 22¡ÆC and after five second warm-up unless otherwise specified. |
Offset |
-25 |
0 |
25 |
mmHg |
Zero |
Sensitivity |
4.95 |
5 |
5.05 |
µ/V/V/mmHg |
¡¡ |
Calibration |
97.5 |
100 |
102.5 |
mmHg |
Output of sensor with no pressure applied and a 150 K¥Ø resistor shorted across + IN to + OUT |
Symmetry |
¡¡ |
¡¡ |
5 |
% |
¡¡ |
Accuracy (-30 to 50 mmHg) |
-1 mmHg -1% of Reading |
0 |
+1mmHg +1% of Reading |
¡¡ |
Combined effect of sensitivity, repeatability, nonlinearity and hysteresis errors. Per AAMI BP22 |
Accuracy (50 to 300 mmHg) |
-3 |
0 |
3 |
% output |
Combined effect of sensitivity, repeatability, nonlinearity and hysteresis errors. Per AAMI BP22 |
Temp Coeff – Zero |
-0.3 |
0 |
0.3 |
mmHg/¡ÆC |
Over a temperature range of 15¡ÆC to 40¡ÆC. |
Temp Coeff – Sensitivity |
-0.1 |
0 |
0.1 |
%/¡ÆC |
Over a temperature range of 15¡ÆC to 40¡ÆC. |
Frequency Response |
1,200 |
¡¡ |
¡¡ |
Hz |
Previously qualified, not tested in production |
Phase Shift |
¡¡ |
¡¡ |
5 |
degrees |
Previously qualified, not tested in production |
Offset Drift |
¡¡ |
¡¡ |
1 |
mmHg/8 hrs |
Normalized offset/bridge voltage – 8 hours after 20 second warm-up, Previously qualified, not tested in production
|
Merit Sensor Systems Çѱ¹ ÃÑÆÇ (ÁÖ)´ÙÀϾÆÀ̺ñ¾¾